Using air compressor to calibrate sensor

Hello, right now i am calibrating a pressuresensor and im using this aircompressor for the calibration:
6-4 - JUN-AIR

The measure span is from 0-7 psi, and the accuracy has to be +-0.15 psi.

But the questions is, how accurate is the pressure gauge on this aircompressor?
I tried to search on the internet, but i couldn't find the accuracy any places.

This is the pressure gauge:
5130000 - Pressure Gauge (0-16 bar) - Gast Manufacturing
 
  #2  
Old 03-02-16, 08:55 AM
P
Group Moderator
Join Date: Mar 2003
Location: NC, USA
Posts: 25,960
Received 1,766 Upvotes on 1,579 Posts
The gauge that comes with any compressor is technically not calibrated and can not be relied on to calibrate other gauges. And, you want to measure 0-7 psi within +- .15 psi but then link to a gauge that goes to 230psi which is also not calibrated???
 
  #3  
Old 03-02-16, 08:58 AM
C
Member
Join Date: May 2015
Location: USA
Posts: 2,982
Received 108 Upvotes on 87 Posts
At best it will be +/- 1 % of full scale, and probably worse since it's basically a spring gauge. And with a full scale of 230 PSI, you can barely make out where 7 PSI is on the gauge.

No where near accurate or precise enough for what you are trying to do. I'm a afraid you will need a much more expensive instrument to achieve your precision and accuracy.
 
  #4  
Old 03-03-16, 03:06 AM
GregH's Avatar
Super Moderator
Join Date: Jan 2001
Location: Manitoba
Posts: 9,498
Received 66 Upvotes on 61 Posts
The style of gauge you linked to is nowhere what you need for the accuracy you said you want.
A digital gauge would be needed.

What exactly is the sensor you are trying to calibrate measuring and does the gauge need to be certified.
If what you are calibrating is a critical component for something you would do well to have a gauge with NIST certification for accuracy.
We don't know if you are working with an engine sensor or a medical device or a missile.

Not sure if certification is avaiable for this gauge but is the type you need.
 
  #5  
Old 03-03-16, 08:55 AM
G
Member
Join Date: Nov 2007
Location: MI
Posts: 2,357
Upvotes: 0
Received 1 Upvote on 1 Post
At 7psi an accuracy tolerance of ±.15 is 2% of span so it's critical you get a gauge with a full scale in that range in order to maintain accuracy. Damaging such a sensitive gauge with a 120psi compressor is a very real possibility. It would be preferable to use a small hand pump like a basketball inflator or a Mity Vac if you could plumb it to a regulator that's capable of maintaining pressure that low.

A 230psi gauge has no reliable accuracy down at 7psi.
 
  #6  
Old 03-03-16, 11:18 AM
GregH's Avatar
Super Moderator
Join Date: Jan 2001
Location: Manitoba
Posts: 9,498
Received 66 Upvotes on 61 Posts
There is a bit of a difference between gauge accuracy range when comparing digital to analog
gauges.

Within reason a digital gauge has an accuracy that remains constant through out its range.
If you are looking to calibrate at 7 psi you would not want to do it on a 10,000 psi gauge but a 300 psi gauge should not be a problem.
Depending on how sensitive the object of your calibration is a gauge that has a NIST certification is guaranteed to read correctly and if you do not need that certification to purchase the same gauge without the certificate is often 1/3 cheaper.
I use a NIST certified digital gauge that has a range of 0 - 150 psi and calibrate a medical device at 3 - 20 psi.

If we knew what is being calibrated we may be able to help more.
 
 